Here the responses that
we have received:
5/31/99 - Dr.H. Kiefer
( Germany )
Dr. Dr Olavo,
Your photoes came out very
fine. Unfortunately, we have this problem
very often, since we
have to revise hundreds of hips implanted from 1975 up
to now.
I would do this hip
in my common procedure with a cementless technique:
For the acetabulum,
i would use a reinforcement ring (Schneider-Burch,
from Aesculap or Sulzer
Companies), with bone grafting and cancellous screws.
For the stem, Iwould
use a long Bicontact revision stem (300, 340 or 380mm
long, Aesculap) via
a transfemoral or dorsal approach, gone grafting and
distal interlocking.
I copied you some pages from the 1998 edition of the
Bicontact-book, whre
I wrote a chapter dealing with preop. planning. I
hope, you can open
and read it for transmission quality.
Maybe, you can get
a copy of this booklet from Cid.
I'm sorry, that
the videoconference via netmeeting will not work, as,
up to now, I do not
fave a camera and a microfone for my computer. I think I
will have to get these
things within the next weeks.
So we would have to
telephone at 5 h Brasil time.
Hartmuth Kiefer
Dear Dr Olavo,
I'm sorry that it did't
work. I sent you 3 e-mails with additional scans
with examples how to
operate on your special case.
May be you try again
to call me.
Yours
Hartmuth
5/31/99 Dr Pedro
Tucci - Brazil
JOlavo,
Parece que na radiografia
de 1192 já havia sinal de soltura do acetábulo
e mal posicionamento do
componente femoral.
Em 0299 houve fratura
da diáfise com soltura do acetábulo.
Que tipo de papo foi usado
para convencer o paciente que poderia esperar
até 0599 para fazer
algo?
Enfim, estamos diante
de uma caca total, para a qual só resta rezar e,
se o paciente for suficientemente
saudável, tentar refazer o acetábulo
com enxerto e reforço
de metal, com enxertia e colocação de um
componente femoral de
tipo longo e não convencional, torcendo para
integração
do enxerto e não-infeção.
Em caso muita dor e impossibilidade
de reoperação segura, a resseção
será a via final,
indesejável.
Um abraço.
PTucci
5/31/99 Dr. H. Kiefer ( Germany
)
To your case:
For the cup I would use an
acetabular reinforcement ring type Schneider-Burch
from Aesculap or Sulzer
Companies.This enables you to upgrade the acetabulum
with bone grafts and allows
partial weight bearing from the beginning. Only
the interface between
reinforcement ring and PE-cup is cemented, so, in total
it is a socalled cementless
anchorage. If you would use a primary cemented
cup, you will get the
next loosening within very few years.
This is also true for
the cemented stem. Because of the large shaft defect and
the very thin remnants
of cortical bone you should try to recover the bone
stock using bone grafts
in combination with a cementless "ingrowing" long
revision stem and distal
interlocking (Bicontact revision stem, Aesculap).
Again this allows partial
weight bearing from the beginning for 10-12 weeks. I
have done many of these
procedures with very good success. My favorite approch
in these cases with a
horizontal semicircular osteotomy of the shaft at the
level of the fracture
(about 3 cm above the tip of the stem).
Cid shall give you the
Bicontact booklet from 1998, were examples of this
technique are shown.
Sorry again for our bad communication.
If you like, you may call me again on
tuesday, 1.June at home
between 5 and 6 h p.m. Brasil time.
Best regards
Hartmuth Kiefer
6/01/99 Dr. Bartha Lajos
( Hungria - Budapeste )
Bonjour Cher Ami,
I faut faire une reprise
de prothese et changer pour une prothese de l anche tumeur avec
une tige fémorale long, n importe quel production il y
a plusieur. J ai montré pour notre spécialiste
et on a dit ceci. Jéspaire tu vas bien. Je vais bien installé
encore une fois mon programme NetMeeting. J éspaire nous
pouvouns parler cette été. Ton fiston vas
bien ? Tudo bem? Obrigado, je t´embrasse
Lajos
|